期刊論文
學年 | 90 |
---|---|
學期 | 1 |
出版(發表)日期 | 2001-10-01 |
作品名稱 | A reevaluation and discussion on the threshold limit for hemolysis in a turbulent shear flow |
作品名稱(其他語言) | |
著者 | 盧博堅; Lu, Po-chien; Lai, Ho-cheng; 劉佳興; Liu, Jia-Shing |
單位 | 淡江大學水資源及環境工程學系 |
出版者 | Elsevier |
著錄名稱、卷期、頁數 | Journal of biomechanics 34(10), pp.1361-1364 |
摘要 | One major factor of red blood cell damage induced by artificial heart valves is the magnitude of turbulent shear stresses in the flow field. An often-cited threshold for hemolysis is 400 N/m2 (Sallam and Hwang, Biorheology 21 (1984) 783). This value, however, was measured with a one-component laser Doppler anemometer and was not calculated from the major principal Reynolds shear stress, therefore underestimatng the threshold. This study performed flow field measurements under similar conditions, using a two-component laser Doppler anemometer and determining major principal Reynolds shear stress, to reevaluate the hemolytic threshold as 800 N/m2 with an exposure time of 1 ms. The Kolmogorov length scales, approximately 9 μm, are capable of inflicting direct damage to the red blood cells. These results will serve as a more accurate reference in future heart valve design and testing. |
關鍵字 | |
語言 | en |
ISSN | 0021-9290 |
期刊性質 | 國外 |
收錄於 | |
產學合作 | |
通訊作者 | |
審稿制度 | 否 |
國別 | TWN |
公開徵稿 | |
出版型式 | ,電子版 |
相關連結 |
機構典藏連結 ( http://tkuir.lib.tku.edu.tw:8080/dspace/handle/987654321/44733 ) |