Improvement of Pulse Amperometric Detection Integrated Automated Flow Injection Analysis of Ethylenethiourea Determination
學年 93
學期 1
出版(發表)日期 2004-12-01
作品名稱 Improvement of Pulse Amperometric Detection Integrated Automated Flow Injection Analysis of Ethylenethiourea Determination
作品名稱(其他語言)
著者 林孟山;王俊勝;陳佩彥;葉秀炫
單位 淡江大學化學學系
出版者 臺北市:中國化學學會
著錄名稱、卷期、頁數 Journal of the Chinese Chemical Society=中國化學會會誌 51(6), pp.1293-1296
摘要 Improvement of pulse amperometric detection (PAD) method is demonstrated in determination of ethylenethiourea (imidazolidine-2-thione, ETU). The anodic detection of ETU will produce polymeric film on an electrode leading to an inactive electrode surface. Here, the PAD method was used to remove the polymeric film formed on the electrode surface between ETU detection. Further, the scheme was integrated with automated flow injection analysis (AFIA) for determining ETU. The operational parameters of PAD in the AFIA system were discussed thoroughly. The analytical characteristics of the system were evaluated at optimum conditions. The linear range of calibration plot was between 20 to 300 μM (the correlative coefficient, r = 0.999) and the detection limit was 0.9 μM (S/N = 3). The relative standard deviations of detection of 50μM ETU were 0.82% with and 9.07% without PAD scheme. The results indicate the system is a very promising tool for ETU determination. Finally, the matrix effects of two water samples that were collected from a campus and a farm show good recoveries of 92% and 96%.
關鍵字
語言 en
ISSN 0009-4536
期刊性質 國內
收錄於
產學合作
通訊作者
審稿制度
國別 TWN
公開徵稿
出版型式 ,電子版,紙本
相關連結